Leakage Test Diagrams; Test #1 - GE T2100 Service Manual

Hide thumbs Also See for T2100:
Table of Contents

Advertisement

1
Ground-wire-leakage-to-ground
2
Chassis-leakage-to-ground

Leakage Test Diagrams

Test #1

Ground-Wire-Leakage-to-Ground
Line
Neutral
Gnd
Revision A
Installation: Equipment Assembly
WARNING
Total system leakage current must not exceed 300
microamperes.
Table 1. Leakage Tests and Maximum Allowable Leakage Currents
Test
These diagrams show only a representation of how a typical leakage
current tester functions. Follow the instructions provided with the
leakage current tester that you use.
Tester
Tester
Polarity
power
Norm
cord
Meter
connectors
GE T2100 Treadmill
2021403-031
Applies To
GE T2100 Treadmill
GE T2100 Treadmill
"To be tested" power connector on back of
tester (may not be labeled on some testers).
UUT
power
Line
cord
Neutral
Rvs
Gnd
1K
V
Maximum Current (µA)
Gnd open = 1000
Gnd open = 500
Gnd closed = 100
Unit
under
test
(UUT)
016
3-13