GE T2100 Service Manual page 91

Hide thumbs Also See for T2100:
Table of Contents

Advertisement

TDU Event /
Error Code
0x09
0x0a
0x0b
2021403-120F
Status
Definition and
Detail Description
Warning
Lower elevation limit
reached:
• <00> Not an error
– occurred during
calibration.
• <01> Error –
occurred during
normal operation.
Warning
Upper elevation limit
reached:
• <00> Not an error
– first occurrence
during calibration.
• <01> Not an
error – second
occurrence during
calibration.
• <02> Error –
occurred during
normal operation.
Emergency stop was
Warning
initiated:
• <00> Stop
command
initiated from
CASE.
• <255> Stop
command
initiated from the
ESTOP button
T2100 Treadmill
Troubleshooting
Check These Items
Indicates that the elevation
system extended beyond the
upper elevation limit.
1.
Powercycle the T2100
Treadmill and monitor
the T2100 Treadmill for
reoccurrence of this error.
(This may be an isolated
incident.)
2.
Initiate the self-calibration
feature.
3.
Check the elevation
potentiometer. See
"Elevation Potentiometer
Replacement" on page
122.
4.
Replace the elevation
potentiometer.
Indicates that the elevation
system extended beyond the
upper elevation limit.
1.
Powercycle the T2100
Treadmill and monitor
the T2100 Treadmill for
reoccurrence of this error.
(This may be an isolated
incident.)
2.
Initiate the self-calibration
feature.
3.
Check the elevation
potentiometer. See
"Elevation Potentiometer
Replacement" on page
122.
4.
Replace the elevation
potentiometer.
Emergency stop is engaged
(or has been engaged). May
be useful in determining if
a "sudden stop" occurred
because the emergency switch
was engaged.
Disengage the emergency stop.
91