Leakage Tests; Shroud Removal - GE T2100 Service Manual

Hide thumbs Also See for T2100:
Table of Contents

Advertisement

FRU Replacement

Leakage Tests

GE Healthcare recommends that you perform the electrical leakage tests whenever
internal assemblies are serviced.
See
"Performing Leakage Tests" on page

Shroud Removal

To remove the T2100 Treadmill shroud for the internal visual inspection or for service:
• Disconnect the power cord from the rear connector panel.
• Remove the screws on the shroud back and remove the back panel.
• Remove the 2 bolts on each side of the shroud.
• Lift off the shroud.
Item
1
2
106
Description
Back Panel
Shroud Side Bolts
T2100 Treadmill
47.